OK - a few random thoughts..
Hi
@WobblyHand - a newer fellow curious about what is going on here, welcome!
Although I originated the sentiment in the title,
@homebrewed is the relative mainspring. I had not considered that moderators might want to rule our explorations out of order for being somehow non "hobby-machinist", so in our defence, I cite the motivations.
Having a chunk of steel, likely acquired via some route other than purchased from a source with the composition guaranteed, is a problem!
One might want to know if it is potentially hardenable, or maybe free-machining by having some lead in it. We really do want to know if the bit of cast iron is semi-steel, or hardly better than pig-iron weights. We would like to know if the stuff is a heat-hardenable alloy, or a carbon steel, and so on.
Doing something like this has me had me thinking about how one might even turn or drill lead, and how to fabricate the enclosure. Everything about trying to make this gadget in a way that lots of HM members might manage to get together is about something we believe they would find useful. On the way, we become educated in some practical nuclear physics, right down to the numbers. We have weeded out the wishy-washy stuff. The thread is now hardball about the science, and we have appreciated and set out what it takes to get this data.
I do agree the thread is huge, and it would take someone with more than average perseverance and interest to trawl the whole thing, but that journey details our learning curve. At this stage, we freely discuss atomic absorption spectra with all that already under our belts. We should perhaps pause a bit, and periodically throw in summaries, and potted explanations, so new readers not so immersed in it, can take advantage.
I hesitate to try for "Needing more than a spark test (2)". That's hiking the generation number like a trash Hollywood sequel, but it would at least re-start the number of pages.
-------
Thinking about the energies we want to detect, there is the question - can we reliably detect lead? Can we even detect Carbon? Nice to know, because that is such an important ingredient. It's presence is most of what allows the spark test to work. From where we started, with the pocket-Geiger, we have the excerpts of energies in a couple of PDFs.
Lead is atomic number 82. Check it out on the second PDF, but also display the first, to get at the top line column titles.
There is no chance a incoming 60KeV gamma can get the K-shell electrons in lead (Pb) to emit anything. 74.9KeV is just too much, and 84.9KeV is worse!
But look, the L-shell electrons need only 10.5KeV, and 12.6KeV to shift, and that is right where our detector can work! See those two together, and you suspect lead.
So ask, can one tell if it is contaminated by Rhenium, Osmium, Iridium, Platinum, or Gold? Sure you can. Consider first the likelihoods. Then consider the resolution of our proposed gadget. Much of the design philosophy I went for was not to settle for a smudged pulse. It needed to be good enough, and have a measurement accurate enough, to try and separate these.
Think through all the common metals and stuff in our steels. Cr, Mo, S, Mn, Fe, Zn, and so on. It's clear that not-very-good analysis, with poorer resolution might persuade is that what we thought was lead might be sulphur, This is why we need the bucket statistics to be
smart! Also, the resolution as high as we can get it. A lot of the software around struggles with getting around stuff like baseline shift, phase delay, pulse stretching and the like. My approach is a bit hardball. DC coupled, or clamped, with high enough bandwidth and low enough noise to make such "smart" guesswork unnecessary, if I can make it so.
One thing I had not considered is how little of our steels are all from ore. Except for perhaps a Katana blade, or special purpose steels, recycled scrap in the mix now has become an alloy with a proportion from all sorts of steel, and has been becoming steadily more radioactive. So long as it's not noise, this might even be an advantage. If it makes the steels glow X-Rays, that's OK
I don't expect we will ever be bothered by our steels ever "getting warmer"
Consider the important carbon. Only 6 electrons, and all of them in the K-shell. Only a feeble 227eV will it yield. The sensor we hijack from the pocket-Geiger kit will only have absorption probability around 3% for that. Even so, it is not zero, and if we have a low enough noise floor, we can scale the measure to account for the sensor. Then again - why bother? So long as the count is characteristic of the element in calibration with that sensor, the element is identified. It's bucket is incremented, and the display will show it there, regardless the info came from a small signal.
It's true the axis of the display plot may need to be scaled to account for the sensor curve, or maybe a logarithmic expansion to some base to exaggerate the low levels could be useful, but the key thing is - we can perhaps detect carbon. It becomes a thing about very low noise amplifier detection technology, which fortunately these days, is actually reasonably affordable.
As for the processing, it takes more than a little Arduino, but a 55 bucks Raspberry Pi can stomp on it! Very high speed DSP, and suchlike are not needed if one is happy to capture the pulse(s) with high speed analog stuff, and analyze at leisure. It just means you have to wait more seconds for the counts to build up. Mark has actually experimented with how often the scintillations happen.