I have not yet given thought to filter discrimination tricks to mask the unwanteds, but what you say does trigger a thought.
In messing with the geometry, I tried to minimize photon routes from lead shielding surfaces that could arrive at the sensor. If the outer tube is large enough radius, then the photon routes at the diode are shadowed to all except the test piece, and the surface under it. That might be something low density, thick enough, and made of stuff that we can't get a glow from. I was thinking a bit of polystyrene insulation might do. Anything will do, so long as it's all carbon and hydrogen and maybe chlorine (PVC). PTFE is nice. Flourine X-rays are too low to care!
If we do not have the test piece sat on a sheet of lead, the lead peaks go away The diode only sees the back of it's tube lead shield, not direct top surfaces that saw a 59keV photon arrive.
The lead, if excited, is capable of delivering L-shell Lα1 = 10.55keV, and Lβ1 = 12.61keV, which is right in the middle of energies we would collect.
The values for lead are:
For Pb
K shell --> Kα1 = 74.9694 Kβ1 = 84.936
L shell --> Lα1 = 10.5515 Lβ1 = 12.6137
The energy is not high enough to get a glow from the K-shell, but we would see the lead there from L-shell responses. It may not matter, because, out of your list of test metals, only zinc comes close with 9.57keV. If we have preserved our pulse shapes well enough, we should still be able to see the separate peak.
The obvious question is - if we do have lead responses, are there any filtering tricks one can use to let the wanted X-rays build up count, to tip the scales against the lead response?
Going where I have not yet fully understood it yet.
On the getting of calibration pure elements, I suppose I may have to gather some. On calibration, one can have software setups to deliberately spread the X-axis buckets over a more limited range, to "zoom in", using calibration elements nearer to what is suspected to be in the alloy.
e.g. Setting the "zoom" to have more buckets devoted to collecting energies in range 2-8keV, the cadmium L-shell would have handy 3.1keV and 3,3keV responses. We would have ignored the 23.1keV and 26.1keV. Even at the low 2% probability of getting a diode response at 3keV, we may see it if we wait some seconds (minutes??)
Do tell what Amazon search term you used?
In my design, there is now provision for inserting a disc of metal "filter" up the shadow tube to be over the sensor, where Am241 energy cannot directly see it.
In messing with the geometry, I tried to minimize photon routes from lead shielding surfaces that could arrive at the sensor. If the outer tube is large enough radius, then the photon routes at the diode are shadowed to all except the test piece, and the surface under it. That might be something low density, thick enough, and made of stuff that we can't get a glow from. I was thinking a bit of polystyrene insulation might do. Anything will do, so long as it's all carbon and hydrogen and maybe chlorine (PVC). PTFE is nice. Flourine X-rays are too low to care!
If we do not have the test piece sat on a sheet of lead, the lead peaks go away The diode only sees the back of it's tube lead shield, not direct top surfaces that saw a 59keV photon arrive.
The lead, if excited, is capable of delivering L-shell Lα1 = 10.55keV, and Lβ1 = 12.61keV, which is right in the middle of energies we would collect.
The values for lead are:
For Pb
K shell --> Kα1 = 74.9694 Kβ1 = 84.936
L shell --> Lα1 = 10.5515 Lβ1 = 12.6137
The energy is not high enough to get a glow from the K-shell, but we would see the lead there from L-shell responses. It may not matter, because, out of your list of test metals, only zinc comes close with 9.57keV. If we have preserved our pulse shapes well enough, we should still be able to see the separate peak.
The obvious question is - if we do have lead responses, are there any filtering tricks one can use to let the wanted X-rays build up count, to tip the scales against the lead response?
Going where I have not yet fully understood it yet.
On the getting of calibration pure elements, I suppose I may have to gather some. On calibration, one can have software setups to deliberately spread the X-axis buckets over a more limited range, to "zoom in", using calibration elements nearer to what is suspected to be in the alloy.
e.g. Setting the "zoom" to have more buckets devoted to collecting energies in range 2-8keV, the cadmium L-shell would have handy 3.1keV and 3,3keV responses. We would have ignored the 23.1keV and 26.1keV. Even at the low 2% probability of getting a diode response at 3keV, we may see it if we wait some seconds (minutes??)
Do tell what Amazon search term you used?
In my design, there is now provision for inserting a disc of metal "filter" up the shadow tube to be over the sensor, where Am241 energy cannot directly see it.