- Dec 20, 2012
- 9,422
Coomba, this is from Long Island Indicator Service's site:
Test indicators can also be calibrated on a surface plate using certified gage blocks. The indicator is securely fastened to a stand and the contact point is brought in contact with a gage block of a given size. The contact point must be parallel with the surface of the block for most manufacturers. Interapid test indicators are an exception and should be at a 12-degree angle, approximately. The gage block can then be removed and replaced a number of times to check for repeatability. Be certain that discrepancies in repeatability are not due to poorly tightened clamps, flimsy stands or other factors. Usually one quarter of a graduation repeatability is allowable, but check with the manufacturer's calibration specs for your particular model.
Errors in repeatability indicate a need for cleaning and, possibly, repair. Do not attempt this without experience.
Accuracy in travel is checked by replacing the gage block with one larger. Very small intervals are required. Ideally you'd want to check the travel at every half revolution, or better. During this procedure be certain that the gage blocks are properly wrung to each other and to the surface. In general, accuracy should not vary more than one graduation per dial revolution on .0005" indicators.