@RJSakowski :
We agree on the method, though maybe use some terminology differently.
If the plate underneath is level, then the bubble will stay in the same (wrong) place, regardless how you rotate it.
Thus for Case 2: Reversing the level, in my mind, does not change the direction of the bubble, but that of course, is relative to the level.
Relative to the rest of the room, I suppose it does look as if it has changed. The fix is obvious.
Getting to Case 3: Yes indeed. Adjusting to have the level read the same amount off when the level is reversed has fixed the calibration. All it means is you have calibrated correctly, even though using using a line through the surface that was not level. We don't have to use horizontal as the calibration point. We can use something a bit off, because reversing allows us to get there anyway. Having calibrated, most folk will do the obvious, and seek the proper rotation, to see the bubble in the middle, and then reverse it, and expect to see the bubble again end up in the middle.
I suppose most of us have had to tangle with calibrating levels at some point. I put down a 123 block against it, to be able to set the level down again exactly reversed. I will obviously have to be doing more of this, but first, I wanted to show folks what it looked like inside, and how much it cost. At some stage, lacking an autocollimator, I will be walking a level in equal increments along a lathe bed, plotting the ups and downs.
Your point about the little cross level needing to also be correct is interesting. I thought that in a tilted plane, there was only one line through that could be level. The line at 90° to it can hardly also be so. To get both vial to read OK, the surface under had better be set up level beforehand.
We agree on the method, though maybe use some terminology differently.
If the plate underneath is level, then the bubble will stay in the same (wrong) place, regardless how you rotate it.
Thus for Case 2: Reversing the level, in my mind, does not change the direction of the bubble, but that of course, is relative to the level.
Relative to the rest of the room, I suppose it does look as if it has changed. The fix is obvious.
Getting to Case 3: Yes indeed. Adjusting to have the level read the same amount off when the level is reversed has fixed the calibration. All it means is you have calibrated correctly, even though using using a line through the surface that was not level. We don't have to use horizontal as the calibration point. We can use something a bit off, because reversing allows us to get there anyway. Having calibrated, most folk will do the obvious, and seek the proper rotation, to see the bubble in the middle, and then reverse it, and expect to see the bubble again end up in the middle.
I suppose most of us have had to tangle with calibrating levels at some point. I put down a 123 block against it, to be able to set the level down again exactly reversed. I will obviously have to be doing more of this, but first, I wanted to show folks what it looked like inside, and how much it cost. At some stage, lacking an autocollimator, I will be walking a level in equal increments along a lathe bed, plotting the ups and downs.
Your point about the little cross level needing to also be correct is interesting. I thought that in a tilted plane, there was only one line through that could be level. The line at 90° to it can hardly also be so. To get both vial to read OK, the surface under had better be set up level beforehand.