RWW,
1. find a flat surface, mill table, vice, etc the length of your level. I use my band saw table. A circular saw table OK.
2. set the left end on your left hand. This end is not adjusted, the one with the cross vial.
3. Read the left end of the bubble. The surface must be level enough to be able to read one of the graduations, but flat.
4. Remember the location measured from the hash mark and the direction.
5. Rotate the level 180^. (Now the jams nuts are on the left side. First iteration)
6. slightly loosen the vial case screw on the right.
7. loosen the jam nut on top, now on the left hand.
8. adjust the bottom nut until the bubble takes out HALF the difference. hand tighten jam.
9. RETURN TO 5
Finally, there will be no adjustment.
when satisfied, tighten jam and case screw. You will have to do this several times a year.
Lastly, insert a thick card stock under one end, mine measures .007", and tell me how many graduations and partial graduations this moves the bubble.
Tell me the thickness of your stock. The length of your vial tells me this will be a great level. Now it is adjusted to read "plumb" as it is called.
Mike
Remlap AL